This was sent to me from Animas the other day and I figured that I’d pass it along…
Important Information about the OneTouch® Ping™ Glucose Management System
and the Animas® 2020 Insulin Pump
Dear Animas Pumper:
Animas is dedicated to supporting pumpers with quality customer service and products. As part of that commitment, we reach out whenever we have urgent information about our products.
We have learned of several cases in which pumpers have accidentally delivered excessive insulin when they ignored pump warnings and did not disconnect from their infusion set before loading the pump cartridge. In each of these cases, the pump had been damaged. Because of this damage, the pump delivered the whole cartridge of insulin during the “load cartridge” step of the priming sequence.
Failure to follow safety instructions and disconnect your infusion set from your body during the “rewind, load and prime” steps can lead to unintended delivery of insulin. This can cause serious injury or death.
This warning is in your Owner’s Booklet or User Guide and was part of your pump training. The pump also warns you to disconnect before these steps (see below). NEVER IGNORE THESE WARNINGS.
There is NO SAFETY RISK from this condition if you ALWAYS disconnect your infusion set from your body when:
rewinding the motor
loading the cartridge
priming the infusion set tubing
tightening the cartridge cap
Please note, this issue does NOT occur during basal and bolus insulin delivery.
How Will I Know if My Pump is Affected?
If you see a pump warning message that says: “No Cartridge Detected. Delivery Disabled” after the “Load Cartridge” step; or
If you see repeated “no prime” warnings (see below) that are not related to an occlusion alarm, auto-off alarm or loose cartridge cap; or
In some cases, you may also notice that the pump’s screen is hard to read or goes blank.
If you notice any of these things, contact Animas Customer Support at 1-866-949-1525 or call the toll free number on the back of the pump. As always, if you need immediate medical attention call your healthcare provider or 911.
What do I do now?
The key things to remember are:
Follow your Owner’s Booklet or User Guide instructions
Pay attention to all pump warnings
NEVER rewind the pump motor, load the insulin cartridge, or prime your infusion set while connected
Should you have any questions or concerns, please call the Animas Customer Support Department. A trained healthcare professional will take your call at 1-866-949-1525 or you can call the toll free number on the back of your pump at 1-877-937-7867. If you are travelling outside of the U.S., please call 610-644-8990 and press option 1.
To be sure that you receive this important information, we have sent a letter to you via the US Postal Service and a separate letter to your healthcare provider to inform them of this issue.
Sincerely,
Animas Corporation
200 Lawrence Dr.
West Chester, PA 19380