Medtronic 670 G approved 6 months early

Medtronic MiniMed 670G
Medtronic MiniMed 670G Trial Results: 44% Reduction in Hypoglycemia, 0.5% A1c Improvement
Highlights from the first study of a commercial automated insulin delivery device. Plus, Medtronic’s latest software updates!

At the recent ADA Scientific Sessions, Medtronic presented data from its pivotal trial of the MiniMed 670G/Enlite 3 hybrid closed loop system, designed to provide evidence for regulatory approval. Compared to a two-week, “open-loop” phase (pump + CGM without automation), adolescents and adults spending three months on the hybrid closed loop system saw:

A 0.5% reduction in A1c, bringing patients from a low initial A1c of 7.4% to 6.9%;

A 44% reduction in time spent with low blood glucose (under 70 mg/dl);

A 40% decline in time spent in dangerous hypoglycemia (under 50 mg/dl);

An 11% decline in time spent over 180 mg/dl and an 8% improvement in time-in-range (71-180 mg/dl).

This dual reduction in A1c and hypoglycemia demonstrates the power of automating insulin delivery – shaving off lows AND highs, particularly overnight. Take note of the images below, which show the combined daily glucose profiles in adults (B) and adolescents (C) for the open loop run-in (gray) and hybrid closed-loop with the MiniMed 670G (pink). The 670G tightened the range of glucose values throughout the entire day (particularly in adolescents) and was very effective overnight in both populations.

Medtronic has submitted this system to the FDA as planned, with a US launch previously expected by April 2017. As we previously reported, of the 124 patients in the trial, a striking 80% have opted to continue using the device through the FDA’s continued access program – this demonstrates encouraging participant enthusiasm for the system, plus FDA comfort with real-world use of the device.

Additionally, the Enlite 3 sensor looks much better than Medtronic’s current sensor. Its average error in this pivotal trial relative to lab measurement was 10.3%, needing a minimum of two fingerstick calibrations per day. This is slightly worse than Dexcom’s G5 at ~9% with two calibrations per day, but much better than the previous Enlite at ~14%. Abbott’s FreeStyle Libre has an ~11% error, but it does not require any fingerstick calibrations.

This represents the first pivotal trial of a fully integrated commercial product to automate insulin delivery, and it shows what a difference first-generation hybrid closed loop systems can make, even in patients doing pretty well already. Patients were already doing well at study start, though the 670G brought A1c down a further 0.5% and cut hypoglycemia by 44% – it’s rare to see both improve with a diabetes drug or device, and the results were consistent in adolescents and adults. This study enrolled current pump users, some of whom were already on a CGM; it’s not clear how people on injections would like this device. Our hope is that the 670G can really improve outcomes for those with higher A1cs (i.e., over 10%) or at high risk of severe hypoglycemia – both groups were excluded from this study. We also hope to hear more on usability of the system, as some have told us it does require a learning curve.

Notably, this wasn’t a randomized study with a parallel control group, meaning some “study effect” could be present – i.e., patients did better simply because they were in the study. Medtronic chose this design deliberately, as it wanted to complete the study as fast as possible and submit the device to the FDA.

Managing expectations will be key with the MiniMed 670G, as the system won’t do everything, and the algorithm may be conservative for some users (i.e., it may target higher average blood glucose levels than some people may prefer). How does the algorithm work? Based on the Enlite 3 CGM reading every five minutes, the pump software automatically increases or decreases basal insulin delivery to target a blood glucose of 120 mg/dl. It’s a “hybrid closed loop” because a user still needs to bolus for meals and notify the system of exercise. The 670G will NOT give automatic correction boluses based on CGM readings; that means if a meal bolus is missed, it will increase basal insulin up to a certain point to bring blood sugar down, but it won’t give a large automatic correction bolus at one time on its own (future systems will add that). Still, the automated basal insulin delivery takes care of insulin dosing in the background, which is particularly valuable at night (~80%+ time-in-range), and can help mitigate many highs and lows during the day.

Unlike many academic systems that run the closed-loop control algorithm on a smartphone, the 670G algorithm is fully integrated within the pump itself. A patient only needs to wear the Enlite 3 CGM sensor and the MiniMed 670G pump – no need to carry a separate CGM receiver or phone.

In addition to these study results, Medtronic announced some important diabetes data updates at ADA, including information on its next-gen CareLink Pro reports for providers (to help optimize pump settings), its partnerships with Glooko and IBM Watson (both launching this summer), and a new food app with Nutrino (beta app available). See below for more details on these and for additional information on the study.

Adults on the MiniMed 670G (pink) vs. Open Loop (gray). The bands depict the range of blood glucose values experienced across 24 hours in the two study periods. The solid (670G) and dashed lines (open loop) indicate the average glucose at that time point. Adults saw a slightly higher average glucose on the 670G during the day, but a narrower range overall.

Adolescents on the MiniMed 670G (pink) vs. Open Loop (gray). The range of glucose values was much narrower overall at all time of day, particularly after breakfast. Unlike the adults, the adolescents saw a lower average glucose on the 670G at many times of day.

Medtronic 670G Pivotal Trial – FAQ
Q: What impact did the hybrid closed loop have on A1c?

A: The MiniMed 670G drove a 0.5% reduction in A1c from a low initial value of 7.4%, including a 1% A1c reduction in participants with a starting A1c of over 7.5%. The 1% improvement in those with a starting A1c above 7.5% is particularly encouraging, as it shows the potential benefit insulin automation can have on those with higher A1cs relative to current open-loop therapy.

By the end of the study, 55% of trial participants had an A1c under 7%, up from 31% at the study’s start. Inversely, that means that 45% of participants ended the six-month study with A1cs still above 7%. How is that possible? Hybrid closed loop systems won’t be perfect, since they only modulate basal insulin and still require accurate carb counting and meal boluses. If a patient misses a mealtime bolus, for instance, blood sugars will likely stay elevated for some period of time before the higher basal rate brings it back down. The next-gen version is expected to add automatic correction boluses.

Q: What impact did the hybrid closed loop have on “time-in-range,” hypoglycemia, and hyperglycemia?

A: Time-in-range (the percent of time spent between 71-180 mg/dl) improved from 67% during the two-week run-in to 72% during the study, with time under 70 mg/dl nearly halved (6% to 3%), time under 50 mg/dl declining 40% (1% to 0.6%), and time over 180 mg/dl improving moderately (27% to 25%). The results were fairly consistent across age groups, though adults saw more benefit in hypoglycemia reduction, whereas adolescents saw more benefit in reducing highs.

While the time-in-range increase may seem minor, considering the meaningful hypoglycemia reduction and the high baseline time-in-range, these results are still notable. Moreover, the hybrid closed loop system significantly reduced severe highs and lows, tightening the range of glucose values throughout the day (see above). Reducing severe highs (e.g., over 300 mg/dl) to moderate highs (e.g., 200 mg/dl) is clearly a good thing, though it won’t show up as a benefit in 70-180 mg/dl. Consistent with earlier trials, the MiniMed 670G is most impactful overnight, driving participants down to ~140 mg/dl by early morning, tightening the range of overnight values, and eliminating hypoglycemia.

Q: What was the impact on insulin dose requirements and weight gain?

A: The MiniMed 670G increased total daily dose ~7% from baseline (48 units to 51 units), though a smaller proportion of insulin was given as basal insulin. Since the MiniMed 670G only automates basal insulin delivery, the implication is that participants were taking more manual boluses on hybrid closed loop (either a larger number of doses or larger dose sizes). It is possible participants ate differently on the hybrid closed loop, which could also explain the mild weight gain seen with the system (about three pounds in adults and two pounds in adolescents).

Q: How big is the study? How was the study designed?

A: 124 participants with type 1 diabetes (94 adults, 30 adolescents) were enrolled in the study. Adolescent participants had an average age of 17 years and A1c of 7.7%. Adult participants had an average age of 45 years and an A1c of 7.3%.

Participants began with a two-week “open-loop” phase during which they wore the MiniMed 670G pump + Enlite 3 CGM, but without any insulin automation. This served as the study’s baseline. After two weeks, all participants then moved to a three-month, “hybrid closed loop” phase. During this phase, participants used the system unsupervised, meaning these three months reflected “real-world” experience with the hybrid closed loop system. A one-week hotel study occurred in the middle of the 670G period, allowing a comparison between the Enlite 3 and a laboratory reference.

This is a Fully Closed Loop Automated Basal 24/7. The FDA already approved the one that just suspends basal lows 2 years ago. They started with baby steps to ease the FDA into approving this one. The next step will be automated closed loop bolus added! Fully automated and when quicker liquid insulins come out that leave the system faster, this will probably even work that much better. The website shows that people on the non-automated pump improved their HbA1C by .5% with basal automation to lower it to 6.9%. My guess is that rigid trial procedures put a cap on the best effects. I’m willing to bet on the streets (not in the trials) type 1 diabetics will be able to get their HbA1C down to 6.0% with this! Also notice the trial results show Medtronics new CGM is only 10% error from Lab compared to 9% error with Dexcom. So Dexcom is no longer the King of CGM accuracy. Just 1% off from Dexcom.

Are you an employee of Medtronic?

No doubt life changing for folks who struggle to keep their a1c in a range less than say 7%. Question is though is it life changing for someone who hovers around 6% or below? As @Sam19 points out, if those folks can achieve a similar a1c while taking out the mental burden that comes with a lower a1c, then maybe I could be convinced.

There is ZERO doubt that this is a great step forward and I would assume just the beginning of automating diabetes care.

1 Like

No… I’m not an employee of Medtronic. Just a T1D diabetic thats been waiting 27 years for those slow basrards to come out with an Automated Insulin Pump! Waiting 27 years is why I’m being excited about. 27 years with an HbA1C of 8.0% or higher. I’m lucky to not have any complications… but I’m really feeling the push to get my HbA1C down to 6.0% so I can escape complications before the Cure. It would be a shame if a cure comes out in 5 years and I made it Scott Free so long and go down right before a cure! I thought automated insulin pumps would have came out long time ago. I think its somewhat too late because by the time it gets rolling next year… Smart Insulins will be hitting the market soon. Any diabetic would choose a smart insulin over an automated pump, to not have to wear a device, patches, and tubes! So there you go… would a Metronic Rep say this?.. “I hope Medtronics pumps go out of business with Smart Insulins coming out”. May the best technology win!

Mike… that’s not the question Medtronic is asking because 95% of T1D’s can’t get their HbA1C down to 6.0% or below. Congratulations to those few folks like Sam that can get their HbA1C under 6.0%. There’s no need for him to be on here. He can go and live his life complications free. But for the rest of us, that is not the question to ask. The fact the Medtronic trial showed 1.0% improvement with T1D’s already on a manual pump if their HbA1c was above 7.6% before and .5% improvement for those on manual insulin pumps with 7.4% or lower prior to starting the artificial pump. So the study shows that it’s still an improvement for those already with tight control and even more improvement with those above 7.6% HbA1C. Think about it for a second. Those with HbA1C’s at 6.0% don’t have their lives at risk. The rest of us 95% of T1D’s with HbA1C’s way above 6.0% are the ones who’s lives are at risk. So the big question is to save majority of diabetics lives. … not to help you 6.0%'s lower it to 5.5% … it wouldn’t be a life saver for you few Outliers. It might not even be anything special that you few are doing, but your body chemistries are responding better to the type of insulins out there.

1 Like

I’d like a separate thread started where everyone just honestly posts their Average Annual HbA1C for this year. I guarantee you majority are not even close to as low as 6.0%. The few Outliers that can pull it off might be Genetic Freaks that respond unordinarilly well to insulins out there. Jarmir Jagr is still playing NHL Hockey at Age 40 and he said in an interview he feels great and plans to keep playing until age 60. There are genetic freak Outliers even in the diabetic world. Lol Let’s start a thread and everyone post their Average HbA1C for this year.

Feel free to. I suspect in addition to my being a genetic freak, how hard I work towards tight control, how extremely I limit my diet, how much I read up on the subject and effort I put into it is probably significant as well. I tend to think that’s probably the case for others here who are meeting their a1c goals as well. There are a lot of people on this forum with healthy non-diabetic level a1c.

I somehow tend to think it has more to do with a common interest in tight control than a group of genetic freaks just happened to come together here in cyberspace.

2 Likes

I agree with your statement that few T1Ds are able to achieve those sub-6.5 A1Cs, but the data DOES NOT support the notion that all those folks running sub-6.5 or even sub-6 A1Cs can avoid all the complications of the disease. 89 percent of the risk of complications associated with T1D in the DCCT trials was down to something other than A1C, probably some hodgepodge of genetic risk, other inflammation, etc, that we haven’t yet identified. You can dramatically reduce risk if you lower the A1C and the lower the better – but sadly, there are people even on these forums who, for instance, develop gastroparesis, neuropathy or even nephropathy after decades of A1Cs in the low 6’s. So I think even the people who have a low, but still diabetic A1C deserve access to technology that will meaningfully decrease A1C further (and also address other causes of complications, but that’s another story).

1 Like

I’d advise you not to put any money on this A$$umption.

A large number of TuD Forum members have great A1c’s, not because we are “Genetic Freaks” but because we have educated and continue to educate ourselves about all things D and because we use that knowledge to work out butts off in order to obtain those great numbers. I suspect you yourself would have a much better A1c if you put in this sort of effort. If you choose not to, that’s your business. But you are quite a ways out of line when you make comments that imply (or outright state) that we have great A1c’s because we won some genetic D-lottery. Some cases of D are more difficult to manage than others. But I can guarantee you that every single person with a great A1c didn’t achieve that number without a great deal of hard work for which they deserve every ounce of credit they have coming to them.

7 Likes

Thanks @rgcainmd. This is not the first time it’s been stated or implied that I’m somehow the recipient of some tremendous gift for thinking about little else all day every day other than how to best manage my blood sugar. It doesn’t feel like a gift. I work very hard for tight control, I hate that I have to. I didn’t ask for this condition and really hate the assumption that it just must be easier for me than others

5 Likes

Well said!

5 Likes

I think this question of average A1C is really interesting. I’m not really sensing a ton of judgement on either end of this discussion, which is fantastic. I have no data about what the average type 1’s A1C is, but I wonder if the number of very tightly controlled peeps on here seems more common than it might be in the diabetic world, because that group is more likely (I think) to be interested in an online diabetes forum, since diabetes care is a bigger part of their lives. Someone who has very high A1Cs as a result of inattention (as opposed to failed effort, I guess) isn’t probably perusing the forums a ton…I could be wrong, of course.

I feel pretty certain with enough hard work, I could probably get my A1C consistently below 6.0. However, for me personally, the toll is too high. I get extremely anxious and depressed when diabetes care is such a huge part of my life, and I am not convinced the pay-offs are worth it, for me (and I’m talking about the difference between a target A1C of somewhere in the 6s and somewhere in the 5s). This is a choice everyone needs to make for themselves, of course. I just need a balance.

1 Like

Well of course-- this forum is at least to some extent an interest group in tight control. If you go to weight lifting forums you’ll likely find people who can on average lift more weight than average people, etc.

3 Likes

You’ve cut and pasted a large block of text from another website. You did not put that text in “quotes” to alert the reader that these are not your words. (This forum has that function embedded in the composing box. The " icon is how you use it.) Your entry would have been more interesting if you simply pulled some of the narrative from diaTribe and then wrote some of your own comments. A link to the full work is appreciated.

2 Likes

As other responses have noted, we do not consider ourselves genetic freaks.

3 Likes

You don’t quite grasp the concept of peer support, do you??!

3 Likes

What is your financial interest in diabetes? Do you own any pharma stocks, either long or short?

2 Likes

Well, this escalated QUICKLY.

This has been brewing for a while.

2 Likes