I changed to Tresiba a few months back. Was on Lantus. Love how Tresiba has been steady in it’s release etc… Tresiba has been excellent in limiting unexpected lows as well… but…
Last week I opened a new box of Tresiba and opened a new pen. Primed it… also primed it before each shot. For five days I was having crazy high BG readings unlike anything I had ever experienced with long acting insulin. It was as if I was not getting my long insulin. I have been on insulin since 1999 and had never had insulin go bad etc.
So I primed the pen later in the day, dialed up my 40 u dose (on 200u pen) , placed my thumb on the injection button preparing before applying any pressure on the injection button. I wanted to see what was happening with the pen as I would see the dosage window dial slip during injections. I watched as this slight pressure caused the dosage window to wind down by 12 units without expelling any insulin. Then I pressed on the button like I would during an injection and watched far less insulin spill from the pen. I had thought that the insulin within the pen was somehow faulty, but now am concerned that the slightest touch to the injection button was adjusting my dose down after it was set. I repeated this test 5 more times over the sink. Each time the same slippage of the dosage window would wind down, each time the amount of units it would slip on the dosage window was different.
(Diagnosed 1999/Type 1.5/Been using insulin since diagnosis)
Just wondering if anyone else has had this occur/noticed that occasionally the Tresiba seemed off etc. I read some items online about priming issues… but this is different. Also did this with another Tresiba pen/same box… same thing